1. TECHNICAL INFORMATION
1.1 Major features of Sintering furnace CMF-210, CMF-220,
- 10 Steps program patterns
- PID optimized control
- Microwave 1000 watts output
Our typical microwave sintering furnaces operate at a proper GHz frequency with power output in
the range of 1-2 KW. This features,
- Great energy efficiency
- Faster sample heating
- Improved material properties
- Emission reduction
- Wide range of potential applications
The sintering chamber consists of ceramic insulation housing (batch system).
The primary function of the insulation is to preserve the heat generated in the work piece.
The temperatures are monitored by optical sheathed thermocouples placed close to the surface
of the sample.
The system is equipped with stainless steel housing including max. temp. 1,600 deg C, air
circulation, plastic door available and programmable controller with max. 10 program steps, control
output indicator and exhaust device as well.
There are three models for applications of ceramic materials, other industrial needs.
1.2 Technical data
Width : 395mm
Height : 440mm
length : 520mm
Casting : Steel/stainless steel
Firing chamber capacity :
Firing chamber temperature : 1600 deg. C
Chamber operating temp : 1500-1530.C
1.3 Electrical data
Power supply : 220V AC, 60Hz (50HZ option)
Power output : MW 1000 watts
Power input : 1450 watts
2. SAFETY ADVICE FOR INSTALLATION
2.1 The grounded plug with AC power should be connected to the grounding system in power receptacle for prevention of system from any electric shock as well.
2.2 The furnace must not be exposed the direct sunlight.
2.3 Do not place any combustible objects near the furnace.
2.4 Sensor Input/Temperature Measurement
Thermocouple wire should be protected with its shielding tape or meshes for any noise
such as microwave.
3. PRECAUTION FOR SAMPLE CORE INSERTION
3.1 place sample core running with parralel with thermocouples and keep distance(1-2 mm)
not to be close.
.4. FURNACE START-UP PROCEDURE
**** PRECAUTION FOR OPERATION--- this furnace should be
operated under condition below 28℃ ambient temperature and
below 70 % humidity. Air conditioner environment is recommended.
4.1 Insert a plug to the ground system in power receptacle.( refer to 2.1).
4.2 Make sure that the outside cooling fans is turned on.
4.3 Close the door of fire chamber.
4.4 Place the thermocouples suitably, and make sure if the current temperature is indicated at PV and SV
in controller
4.5 Close the outside furnace door. (If the door is opened, furnace would not be operated.)
4.6 Confirm that PV (present value) is equal to SV (setting value) as shown below picture.
fire chamber titimer
6. general setting: this furnace’s original setting value
Setting temperature setting time note
1000 ‘c 16-25min
1200’c 10-15min
1400’c 10-15min
1500’c 10-15min
1500도 sustaining 5-10min
7.Timer setting procedure
7.1 press “side “ key green unit flashing
7.2 press “up key” ” key figures on hour/min/sec, example 2hr 30min 00 sec 2 30 00
7.3 press key for save figures
8. TROUBLE SHOOTING
If you additional assistance is required,
please contact EDEN Ltd.
+82(2) 2254 – 2845 in Korea.